I would suggest it's not that easy, as modern ventilators are exceedingly complicated.
Germany ordered bucket-fulls some time ago, as did one or two other countries, and have allegedly managed to find some extra (Spanish) nurses from somewhere.
For some reason, we seem to be lagging behind many other governments in a variety of ways.
This Financial Times link contains an embedded letter which sheds some light on the more modern kit.
Essentially, as well as the kit, you need the staff who know how to use it.
I have no idea of the authenticity of Dr G's comments, but they do provide valuable explanation, assuming they are valid.
For the record, I am not an NHS ICU consultant in the South East of England… But if I were to imagine having done ICU since 1986 I might make some observations:
The old ventilators were essentially mechanical bag squeezers that inflated the lungs and then paused to let the gas out, before repeating the cycle. They had little sophistication, but meant that a person did not have to physically squeeze the bag, unlike the medical students who were enlisted to do this in 1952 during the polio epidemic in Copenhagen.
**LINK**
In the early 1980s we were ventilating patients after long operations, cardiac operations, head injuries, sepsis and the like. The key point is that the lungs were not the primary cause of the need for intensive care. Indeed, if the lungs started to fail, with Adult Respiratory Distress Syndrome, as it was called then, the patients usually perished. It was during the 1980s that more sophisticated ventilators were developed. We started to get Siemens machines which allowed alterations in the timings of inspiration and exhalation, they also allowed us to put positive end expiratory pressure into play. This meant that the patient would breathe out against resistance. The effect of this is to hold the tiny airways open and the alveoli, and to push fluid back into the circulation instead of having it sitting in the lungs and reducing oxygenation.
These machines are now the norm. They are smothered with sensors and we can measure most things and record changes that allow us to adjust the therapy. For example we can calculate lung stiffness and how it is changing with our therapies. We cannot do this with bag squeezing.
The sensors make the process less dangerous. If we get it wrong, the ventilator can cause barotrauma. At its most extreme we blow a hole in the patient's (diseased) lung and then we have a tension pneumothorax, which is rapidly fatal unless we stab the chest and insert a drain to allow the lung to re-expand. More often, the barotrauma causes microscopic damage that delays or prevents the sick lung from healing. The sensors also alert us if the tubing kinks or disconnects. Unchecked these simple things are quickly fatal.
So, to run a patient with acute lung injury/pneumonia as their primary diagnosis, to keep the oxygen going to the vital organs, we need all the sophisticated settings and sensors and alarms. We also need a highly-trained ITU nurse who can operate the machine, take the readings from it and alert the ITU doctor to adverse changes. When we introduce new ventilators, it takes several weeks for our nurses to become comfortable with them (think airline pilot and new aircraft type). We start them on simple cases where the ventilation settings are basic and the patient is not needing a variety of drugs by infusion and other time-consuming tasks. Once the nurses are happy with the new ventilators, we start to use them on the sicker patients; the nurses are, by now, more reflexive in their use of the ventilators.
So, imagine the perfect storm of non-ITU nurses being put in charge of a desperately sick patient who is attached to a ventilator that none of us has seen before! Mark my words, there are no spare ITU nurses. We will have to start using nurses who have not got the extra qualifications. They will need to be given reliable machines with good ergonomic fail-safe designs.
(I still remember two patients that we killed in the 1980s simply because their ventilators were not properly assembled; the modern machines would alarm and actually describe the fault on their displays.)
Currently all of our machines are German or Swedish. They have undergone years of development so that the alarms give us information according to urgency, the controls are intuitive, every kind of ventilatory mode is possible. They have been tested and CE marked and the companies take responsibility for their machines, through produce liability.
If the government introduces the Johnson Mk 1, or it arranges for a digger company to copy established ventilators, will Matt Hancock assume responsibility for any deaths due to malfunction, or due to our nurses not knowing about the machine's foibles? At present, we may get these machines (possibly) but we would be very unlikely to use them without assurances from the NHS that individual nurses and doctors would bear no responsibility if the machine malfunctioned and destroyed a patient's lungs, or simply failed to deliver 16 breaths per minute 24/7.
Without the nurses we are just window dressing. Sadly, many of our excellent and numerous Spanish nurses have left, because they only accrue seniority pay on their return to the Spanish system if they are working in an EU country. Our loss, Germany's gain apparently.
Bill