      First of all this not a complaint about the supplier or even really Dasqua.

      I just bought a Dasqua 25-50 mm outside digtal micrometer from one of the usual UK suppliers. Overall pleased with the micrometer. It’s not a Mitutoyo but it’a a third of the price.  In the package with the instructions was a calibration certificate.
      Cal-cert-1

      I’m used to looking at calibration certificates and this one did not look or feel (literally) right. Closer examination revealed that the the measured values and claibration “stamp” are halftone printed.

      Cal-cert-2

      Now it’s possible the measurements might be ink jet printed but the font looks like a mechanical type. There is no bleed you would get from inkjet or impression from typing. Also if the calibration values are typed or printed on a blank form why is the serial number hand written in pencil. This is pencil, it’s impressed and there was graphite dust I could wipe off.
      Even more telling is how do you calibrate a 25 to 50mm micrometer with 2.5 to 25 mm gauge blocks
      I’ve seen this before with electronic measuring instruments from the far east. You buy several items and they all have identical errors…
      None of this really matters to hobbyists but if you are an organisation with any sort of quality system or approval it could catch you out.
      I’m not suggesting that the micrometer does not meet it’s specification.Mainly because it doesn’t seem to have one other than “Strictly made in accordance with DIN863” which implies +_ 4 um. Just making some awareness.

      Robert.

      #825889
      JasonB
      Moderator
        @jasonb

        Could they be rubber stamped, that would give a half tone

        #825894
        Bo’sun
        Participant
          @bosun58570

          Isn’t Dasqua an Italian brand, or at least some items of Italian origin?  If so, the seemingly fake certificate doesn’t surprise me.

          #825898
          Robert Atkinson 2
          Participant
            @robertatkinson2
            On JasonB Said:

            Could they be rubber stamped, that would give a half tone

            Hi Jason,

            I’ve never seen a half tone rubber stamp. Is this a new thing? I’d have thought the stamp pressure would have flattened any half tone “bumps”.

            B.T.W I ment to put a question mark at the end of the title. Can you add one for me please?

            Robert

            #825900
            cedric 1
            Participant
              @cedric
              On Bo’sun Said:

              Isn’t Dasqua an Italian brand, or at least some items of Italian origin?  If so, the seemingly fake certificate doesn’t surprise me.

              Italian company originally but their factory has been in Chengdu, China, for many, many years.

               

              #825914
              JasonB
              Moderator
                @jasonb
                On Robert Atkinson 2 Said:
                On JasonB Said:

                Could they be rubber stamped, that would give a half tone

                Hi Jason,

                I’ve never seen a half tone rubber stamp. Is this a new thing? I’d have thought the stamp pressure would have flattened any half tone “bumps”.

                B.T.W I ment to put a question mark at the end of the title. Can you add one for me please?

                Robert

                I was looking at the individual results they look like they may have been stamped, the big circular one is printed

                #825919
                Robert Atkinson 2
                Participant
                  @robertatkinson2

                  I forgot to say, the other discrepancy is no date.

                  Robert.

                  #825924
                  JasonB
                  Moderator
                    @jasonb

                    Probably best to email them an image as get it from the horse’s mouth, look forward to hearing what they have to say

                    #825933
                    old mart
                    Participant
                      @oldmart

                      I have several Dasqua products and they are all better than average quality. I had always suspected that they were not made in Italy because of the price, but it shows that China can make good stuff if they want to. I cannot comment on the certificate as all my Dasquas are at the museum, but I do have my Tesa at home and it has a certificate which has no date on it, but does have the part number and serial number.

                       

                       

                      _IGP3623

                      #826031
                      Robert Atkinson 2
                      Participant
                        @robertatkinson2

                        Jason,

                        I’ve sent dasquatools.com the question. Interestingly their website does not recognise the model number. Additionally dasquatools.co.uk goes to a RGD tools website and Dasqua.co.uk goes to a Cronos website. The plot thickens.
                        I also note that there is no CE or UKCA marking on the instrument, instructions or packaging. This means it’s non-compliant. I’ve asked the supplier about this.

                        Robert.

                        #826034
                        JasonB
                        Moderator
                          @jasonb

                          That I why I linked to the Manufacturer’s contact page not the local distributor

                          #830273
                          DMB
                          Participant
                            @dmb

                            Possibly breaking the law by not having CE or UKCA markings. I wouldn’t blindly trust such markings anyway. Far too much dwindling going on in all spheres today. Best to go by perhaps reputations then general appearance and finish of the product and finally, useage – does it perform to expectations?

                            #830279
                            Nigel Graham 2
                            Participant
                              @nigelgraham2

                              Would a caliper need CE or UKCA accreditation anyway?

                              #830289
                              duncan webster 1
                              Participant
                                @duncanwebster1

                                CE mark means it complies with relevant EU Regulations. I can’t off hand think of one that applies to something which operates on 2 volts or so. UKCA appears to be a dead duck, CE marking continues to be acceptable in IK where applicable.

                                #830294
                                John McCulla
                                Participant
                                  @johnmcculla

                                  I got a Dasqua 25-50mm outside micrometer for Christmas. Here’s the calibration certificate.

                                  The figures are at least different.

                                  PXL_20251228_192603745

                                  #830310
                                  Huub
                                  Participant
                                    @huub

                                    You would expect the measurements and the serial number to be written, not printed. If the results would be entered in a data sheet, than the serial number would also be entered in the same data sheet. So the probability this is fake is pretty high.

                                    My Mitutoyo calibration reports seems authentic.

                                    None of the Chinese reports that came with my tools or machines are somehow trust worthy.

                                    The “test record” of my German Paulimot F207-V milling machine is, based on the measurements (none of the 12 measurement ends on a 0,2,3,4,7 or 9), authentic with a  0.05% probability. The results do not match the actual machine state and the report is 100% printed. That combined it is likely a fake document.

                                    #830330
                                    Nicholas Farr
                                    Participant
                                      @nicholasfarr14254

                                      Hi, I don’t know for sure, but it could be the certificate is printed automatically, and the tester signs it once the test has finished.

                                      Regards Nick.

                                      #830334
                                      cedric 1
                                      Participant
                                        @cedric
                                        On duncan webster 1 Said:

                                        CE mark means it complies with relevant EU Regulations. I can’t off hand think of one that applies to something which operates on 2 volts or so. UKCA appears to be a dead duck, CE marking continues to be acceptable in IK where applicable.

                                        There are two almost identical CE marks. One is EU, the other is Chinese and means nothing. You can look up on the net how to distinguish the subtle differences.

                                         

                                        #830337
                                        cedric 1
                                        Participant
                                          @cedric

                                          CE_Logo-1024x1024

                                          #830350
                                          Robert Atkinson 2
                                          Participant
                                            @robertatkinson2
                                            On duncan webster 1 Said:

                                            CE mark means it complies with relevant EU Regulations. I can’t off hand think of one that applies to something which operates on 2 volts or so. UKCA appears to be a dead duck, CE marking continues to be acceptable in IK where applicable.

                                            Well the  Measuring Instruments Directive 2014/32/EU for a start. And for UK Measuring Instruments Regulations 201.

                                            The EMC directive also applies. The operating voltage (actually 3V for the instrument I bought) is irrelevent to the legislation. There are other directives that apply to items that may not have any electrial components.

                                            Robert.

                                            #830353
                                            Robert Atkinson 2
                                            Participant
                                              @robertatkinson2
                                              On cedric 1 Said:

                                              CE_Logo-1024x1024

                                              On cedric 1 Said:
                                              On duncan webster 1 Said:

                                              CE mark means it complies with relevant EU Regulations. I can’t off hand think of one that applies to something which operates on 2 volts or so. UKCA appears to be a dead duck, CE marking continues to be acceptable in IK where applicable.

                                              There are two almost identical CE marks. One is EU, the other is Chinese and means nothing. You can look up on the net how to distinguish the subtle differences.

                                               

                                              There is no such thing as  a “China Export” mark. The original CE mark was just the two letters but that had legal difficulties becuase you can’t stop people putting letters on their products. So they changed it to a defined symbol which does not match an existing font.

                                              Robert.

                                              #830363
                                              duncan webster 1
                                              Participant
                                                @duncanwebster1

                                                Well being particularly bored I looked up the uk measuring eqpt regs, which are heavily based on the EU directive and I think you’d be stretching a point to apply them to a digital caliper. They are aimed at things like water meters, taximeters etc

                                                Similarly EMC, has anyone ever had interference caused by a digital caliper?

                                                #830372
                                                cedric 1
                                                Participant
                                                  @cedric
                                                  On Robert Atkinson 2 Said:
                                                  On cedric 1 Said:

                                                  CE_Logo-1024x1024

                                                  On cedric 1 Said:
                                                  On duncan webster 1 Said:

                                                  CE mark means it complies with relevant EU Regulations. I can’t off hand think of one that applies to something which operates on 2 volts or so. UKCA appears to be a dead duck, CE marking continues to be acceptable in IK where applicable.

                                                  There are two almost identical CE marks. One is EU, the other is Chinese and means nothing. You can look up on the net how to distinguish the subtle differences.

                                                   

                                                  There is no such thing as  a “China Export” mark. The original CE mark was just the two letters but that had legal difficulties becuase you can’t stop people putting letters on their products. So they changed it to a defined symbol which does not match an existing font.

                                                  Robert.

                                                  Can you not see the difference in letter shape and letter spacing in the above graphic? One is the real CE mark, the other is the Chinese imitation, designed to confuse. Widely used on products I buy directly from China on Aliexpress. Whether it actually stands for Chinese Export or not is moot.

                                                  #830385
                                                  Robert Atkinson 2
                                                  Participant
                                                    @robertatkinson2

                                                    No, It’s just a bad representation. They don’t care about copyright or trademark. I’ve seen otherwise compliant “white goods” from a big manufacturer which had a computer printed label, including serial number, that had a non-compliant CE mark because the printer could not handle the non-standard font.
                                                    Even these labels for sale https://www.labelsonline.co.uk/ce-labels? are non-compliant as the spacing is too close…

                                                    Robert.

                                                    #830396
                                                    Robert Atkinson 2
                                                    Participant
                                                      @robertatkinson2
                                                      On duncan webster 1 Said:

                                                      Well being particularly bored I looked up the uk measuring eqpt regs, which are heavily based on the EU directive and I think you’d be stretching a point to apply them to a digital caliper. They are aimed at things like water meters, taximeters etc

                                                      Similarly EMC, has anyone ever had interference caused by a digital caliper?

                                                      If no-one has had interference from a digital caliper the system must work 🙂

                                                      It is true that very low power items like a digital caliper are unlikely to cause interference. But the EMC directive is not just about causing interference. Items must also be immune to interference. If the result of interference does not cause damage or an unsafe condition the manufacturer can simply state this in their declaration. This approach is used inthe USA for a lot of equipment for compliance with FCC requirements.
                                                      None the less the manufacturer or importer must make the determination of compliance, keep records, declare compliance and mark the item. For a small low power item like a digital caliper this may be just a paperwork exercise. For example complinace with the measuring instrument regulations can be by marking the caliper “not for trade” and incuding information in the instructions.

                                                      Compliance is mandatory, enforcement seems minimal…

                                                      Robert.